Please email phreeqcusers at gmail.com with your name and affiliation to request an account.
Welcome
Guest
Forum Home
Login
Register
PhreeqcUsers Discussion Forum
»
Processes
»
Inverse modelling
»
Phase mole transfers
« previous
next »
Print
Pages: [
1
]
Go Down
Author
Topic: Phase mole transfers (Read 14543 times)
Scarlet
Top Contributor
Posts: 28
Phase mole transfers
«
on:
18/01/18 23:02 »
Hi David,
when I enter these results I find that " Phase mole transfers" is zero,I do not know why
Input
SOLUTION 1
temp 25
pH 7
pe 4
redox pe
units mmol/kgw
density 1
-water 1 # kg
SOLUTION 2
temp 25
pH 9
pe 4
redox pe
units mmol/kgw
density 1
C(4) 16
Ca 8.3
-water 1 # kg
INVERSE_MODELING 1
-solutions 1 2
-uncertainty 1 0.1
-phases
CO2(g)
Calcite
-balances
C(4) 1 0.1
Ca 1 0.1
H(0) 1 0.1
-tolerance 1e-10
-uncertainty_water 1
-mineral_water true
END
output
Solution fractions: Minimum Maximum
Solution 1 9.821e-01 0.000e+00 0.000e+00
Solution 2 1.000e+00 0.000e+00 0.000e+00
Phase mole transfers: Minimum Maximum
CO2(g) 7.263e-03 0.000e+00 0.000e+00 CO2
Calcite 8.737e-03 0.000e+00 0.000e+00 CaCO3
Logged
dlparkhurst
Global Moderator
Posts: 4315
Re: Phase mole transfers
«
Reply #1 on:
18/01/18 23:35 »
The mole transfer is the first number the two zeros that follow are related to the "range" calculation. If you include -range in the INVERSE_MODELING definition, then the minimum and maximum mole transfers are calculated that are consistent with the uncertainties that are assigned. See the manual or the WRR report for more information.
Logged
Scarlet
Top Contributor
Posts: 28
Re: Phase mole transfers
«
Reply #2 on:
29/01/18 10:16 »
thank you mr david, please I have 2 questions:
1- can I know what is the difference between solution and solution spread?
2- what is the role of uncertainty in the inverse modeling?
Logged
dlparkhurst
Global Moderator
Posts: 4315
Re: Phase mole transfers
«
Reply #3 on:
29/01/18 15:20 »
SOLUTION_SPREAD is a spreadsheet-like input for solution data. It is equivalent to SOLUTION, but allows cutting and pasting from a spreadsheet to simplify data input.
You could read the manual or the WRR paper, but uncertainty defines the amount that input concentrations can be adjusted, so it is an estimate of your confidence in the analytical data. The charge imbalances of the solutions are a guide; the uncertainties must be at least large enough to account for the charge imbalance.
Logged
Scarlet
Top Contributor
Posts: 28
Re: Phase mole transfers
«
Reply #4 on:
31/01/18 10:06 »
thank you very much for your answer.
after execution of the program I had these results.
my question is: what can I rely on for an interpretation?
Solution fractions: Minimum Maximum
Solution 1 1.001e+00 0.000e+00 0.000e+00
Solution 2 1.000e+00 0.000e+00 0.000e+00
Phase mole transfers: Minimum Maximum
Calcite -7.895e-03 0.000e+00 0.000e+00 CaCO3
Dolomite 7.895e-03 0.000e+00 0.000e+00 CaMg(CO3)2
CO2(g) -7.895e-03 0.000e+00 0.000e+00 CO2
Pyrite 7.368e-03 0.000e+00 0.000e+00 FeS2
O2(g) 2.763e-02 0.000e+00 0.000e+00 O2
Albite 7.138e-02 0.000e+00 0.000e+00 NaAlSi3O8
Quartz -1.429e-01 0.000e+00 0.000e+00 SiO2
Kaolinite -3.564e-02 0.000e+00 0.000e+00 Al2Si2O5(OH)4
Halite 7.823e-02 0.000e+00 0.000e+00 NaCl
Fe(OH)3(a) -7.227e-03 0.000e+00 0.000e+00 Fe(OH)3
Redox mole transfers:
Fe(3) -7.367e-03
O(0) 5.526e-02
S(-2) 1.474e-02
Sum of residuals (epsilons in documentation): 6.012e+00
Sum of delta/uncertainty limit: 6.011e+00
Maximum fractional error in element concentration: 1.000e+00
Logged
dlparkhurst
Global Moderator
Posts: 4315
Re: Phase mole transfers
«
Reply #5 on:
31/01/18 15:43 »
First, look at your saturation indices.
The mole transfers of silicates is larger than carbonates. Does that make sense? Consider the net effect of the carbonate reactions.
Logged
Scarlet
Top Contributor
Posts: 28
Re: Phase mole transfers
«
Reply #6 on:
07/02/18 11:25 »
I can know by what method can I represent the results? graph or board?
and why we have several results of phase mole transfers and redox mole transfers?
thank you very much
«
Last Edit: 07/02/18 11:59 by Scarlet
»
Logged
dlparkhurst
Global Moderator
Posts: 4315
Re: Phase mole transfers
«
Reply #7 on:
07/02/18 17:16 »
You are on your own to present results.
Mole balance modeling frequently does not give a unique result. Several sets of phases can account for the changes in water composition.
Logged
Scarlet
Top Contributor
Posts: 28
Re: Phase mole transfers
«
Reply #8 on:
14/02/18 15:59 »
Hi David,
please can I do the inverse modeling of these two solutions?
because after execution I find 216 models. Is that right?
SOLUTION 1
temp 25
pH 7
pe 4
redox pe
units ppm
density 1
Ca 21
Cl 9
K 0.8
N(5) 1.6
Na 1.6
S(6) 9.5
-water 1 # kg
SOLUTION 2
temp 25
pH 8.32
pe 4
redox pe
units ppm
density 1
Al 2.499
As 97.05
B 2461
Ba 76.8
Ca 0
Cl 2703
Cs 3.3
Fe 7.661
K 118.37
Li 1160
Mg 187.07
Mn 8.253
N(5) 0
Na 3427.51
Pb 0.021
S 1379.59
U 276.4
Logged
dlparkhurst
Global Moderator
Posts: 4315
Re: Phase mole transfers
«
Reply #9 on:
14/02/18 16:46 »
The number of models depends most on the number of phases and the constraints on the phases. More phases allows more models, and more constraints decrease the number of models.
As for your waters, I suggest you ignore the elements that are not in large concentrations in either the phases or the water. Inverse modeling requires the stoichiometry of the minerals, which may not be well defined for trace elements. So work on the major element (including all major elements in the solution and phases) reactions.
Logged
Scarlet
Top Contributor
Posts: 28
Re: Phase mole transfers
«
Reply #10 on:
15/02/18 15:04 »
thank you very much Mr david.
after eliminating oligo element, I found 12 models.
I want to represent the models on graphs but I do not know what to choose as elements to better represent my results.
Can you help me please?
Logged